The
other day, I was working on a method development project for the USP that
called for low pH. We observed tailing for some of our peaks and speculated
that the lower pH was needed for this reason. To this end, we tried an additive
of 0.5% formic acid in the mobile phase solvents but found a rise in the UV
trace of 200 mAU over the course of our 30 min gradient. In an impurity method,
where low detection levels are of critical importance, this degree of noise
would clearly not be practical. By using TFA instead, however, we were able to
reduce the additive concentration by a factor of five and still obtain the good
peak shapes that we sought. With the lower additive concentration, the trace
slope had risen by about 20 mAU, a mere tenth of the slope obtained previously!
This clearly represented a much more viable option for a prospective USP
method.
You
just have to remember a few things with TFA, though. It can be prone to
oxidation from the atmosphere, so extraneous peaks or a steep UV trace can
result if it degrades. In that case, you would have the same problem as before!
Your best bet is to try single-use ampules of TFA, since these will have
minimal contact time with air. If you use a resealable bulk bottle, you can try
adding a blanket of argon after every time it’s opened to ensure it is kept
free of air. This oxidation is a slow process, so it only becomes an issue
after storage of the opened bottle. Then of course, another thing to keep in
mind about TFA is its MS-incompatibility, as it contributes significantly to
ion suppression. So if you develop a nice method for UV-based analyses with
TFA, just be aware that its applicability will be more limited than with
something like formic acid. Even so, TFA can do wonders for some tailing peaks
of basic compounds, so it is a good tool for the analytical chemist to keep at
the ready.
I hope
these are some helpful tips for you to try in your own method development
process. I was always a fan of TFA for those situations where symmetrical peak
shapes may be difficult to obtain otherwise. The effect of the trace slope can
be one of the more tricky aspects of a UV-based gradient method. And if you
can, maybe also consider using a more gradual solvent gradient. This will have
a direct effect on how steep the resulting UV trace is.
Thanks Josh, very helpful for many of us.
ReplyDelete